Millenium 4 and S/N calculation

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Does any one know how to alter the calculation in millenium 4 so that it calculates the S:N per current USP calculations?

Thanks,
DR
Image

Hi Guys,
Just found a few minutes... For the USP s/n you will need to use 2xPeak Height/noise. The Empower custom field 2*Height*Scale to μV/peak to peak noise used for USP could be applied to Millennium – but unfortunately the Scale to μV field was not available in Millennium. That is where the baseline noise may be more useful as it is measured in mV and can be consistently converted to μV by multiplying by 1000 (as height is measured in μV). Bear in mind the drawback of using baseline noise – it is an average value and is adversely affected by sloping baselines over the extent of the section of baseline selected for measurement. However, peak to peak noise could be used with preset scaling factors.
Once again, hope this helps!
Best regards...
Empower
MVC - Medicon Valley Consult
www.mvc-consult.com
3 posts Page 1 of 1

Who is online

In total there are 35 users online :: 2 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider], Google [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry