- Posts: 11
- Joined: Mon Oct 18, 2010 5:29 pm
I am new to this forum and relatively new to GC operation. Apologies for cross-postings, and to posting this under the wrong heading!
I have an Agilent 6890N GC, with an injector autosampler and combustion furnace, connected to an IRMS. The system is about 6 years old, but has only seen action for about the last 3 years. I have used the GC for only 3 months.
I have been analyzing fatty acid samples (FAME) dissolved in hexane. I did about 2 weeks of analyses before I saw some peak tailing and stopped to perform some maintenance.
After fixing the vacuum on the IRMS, checking the FID, checking the autosampler syringe, cutting bits off the end of the column (~4 years old), putting in a new septum and checking the liner (clean), and going through a software upgrade (MassLynx to IonVantage), the chromatogram I get on a BAME standard is not the same as before maintenance (or on a GC-MS), despite using the same set-up parameters. I have not changed the furnace tube (~4 years old) or changed the Nafion tube (~3 years old).
Differences I've noticed in the BAME chromatogram:
1. Retention times of peaks are slower despite the slightly shorter column. The slow-down is also not linear between peaks between the new analysis and previous analyses.
2. "Profile" of the chromatogram has changed. Compounds that used to produce the tallest peaks are now shorter and vice versa. This presents a problem since my student wishes to use the peak area to calculate absolute concentrations of various C compounds in her unknown samples. If the peaks now have different heights or areas than what we used to get, we can assume that the calculated concentrations will be different too. It's too cost-prohibitive for her to run a second set of samples on a GC-MS to get peak height data, if we could get both peak height and isotope data from the same run for her (these are natural isotopic abundance samples).
The good thing is that, despite the change in peak heights, the C isotope values of the equivalent peaks of the new and old analyses are relatively the same, as are the isotopic trends between peaks.
The problem is not with my IRMS as samples analyzed via another connected peripheral device yields satisfactory results. I don't think this is a software issue either, so it must be a hardware issue with the GC or GC combustion furnace.
I would appreciate any ideas or suggestions or similar experiences anyone may have. Thanks.
Alice