You didn't give a lot of details, but from the description "fast acid column" and the mobile phase, I surmise that you're using a PS-DVB-sulfonate cation exchange resin in the hydrogen form. You didn't give column dimensions or flow rate, so I have no idea about the dead time of the column (i.e., 2.8 minutes may represent no retention at all).
If 2.8 minutes is at or close to the dead time, then this column is unsuitable for your analysis and you will have to go back to "square one" and a more conventional HILIC or reversed-phase column. If 2.8 minutes represents reasonable retention (i.e., if your dead time is less than a minute), then dropping the temperature or increasing the sulfuric acid concentration may help (but no guarantees!).
And, like bhuvfe, I'd be concerned about stability in a low-pH mobile phase.