FID and expected peak areas around LOQ levels
Posted: Fri Oct 22, 2010 12:54 am
Hi guys I am investigating how to demonstrate our planned LOQ of 0.02 %. By the way what is a typical LOQ for a GC/FID impurity analysis of an isocyanate?
Wanting to show >99% purity we are going with area normalisation because every one of the impurities is an unknown compound. No way to check relative responses.
So I have an area of 3,500,000 for the main peak. Then I made a 0.01 % dilution to show suitable signal to noise. But the peak has an area of around 700 which corresponds to 0.02 %, double the size I expected.
A 0.02 % dilution has a peak area of a little over 1400. Am I being unreasonable to expect a peak area closer to 350 for the 0.01 % dilution?
Integration influence from the baseline at these levels is minimal.
Thanks.
Daniel.
Wanting to show >99% purity we are going with area normalisation because every one of the impurities is an unknown compound. No way to check relative responses.
So I have an area of 3,500,000 for the main peak. Then I made a 0.01 % dilution to show suitable signal to noise. But the peak has an area of around 700 which corresponds to 0.02 %, double the size I expected.
A 0.02 % dilution has a peak area of a little over 1400. Am I being unreasonable to expect a peak area closer to 350 for the 0.01 % dilution?
Integration influence from the baseline at these levels is minimal.
Thanks.
Daniel.