Trimyristine reproducibility struggle w/ELSD
Posted: Thu Oct 21, 2010 8:56 pm
Hello everybody,
I have enormous problems achieving reproducibilty - even over a time range of 8 hrs - with my HPLC-ELSD. Sample is trimyristine (saturated triglyceride with 3 C14-chains) in neat THF, concentration about 0,2mg/ml). The peak areas increase up to 30% within a day. I really don´t know what to do anymore and I have put in a decent amount of time to put it mildly.
HPLC-pump: Waters 515
autosampler: Waters 717 plus
detector: Alltech 3300 ELSD
Software: Clarity (Version?)
sample: trimyristine in THF, ca 0,2 mg/ml
column: ODS, 125mm*4mm; 5µm
eluentl: isocratic THF/ACN (45/55)
flow rate: 1,5 ml/min
injection volume: 10µl
run timet: 3 min
ELSD settings:
drift tube temperature: 40C
N2 flow: 1,6 l/min
gain: 1
I´ll give some more details that I think might be asked:
- the lab has an AC
- why 40C drift tube temperature? This is what the manufacturer advises for the given mobile phase, flow rate and sample concentration. Baseline looks awesome.
- The column is not heated, we don´t have an oven.
- All tests you can run are passed excellent according to the manual (i.e. gas flow, dirft tube temperature, clean optics.
- I run a hot cleaning (120C) for 1,5hrs; no changes.
- I took the detector apart and sonicated the nebulizer in MetOH/Water (50/50) . Didn´t change a thing. In the "light trap" I found some fluffy stuff that could be removed easily. Didn´t look like lipid at all.
- The peaks look good. Symmetrical, no fronting or tailing.
- Injection of neat THF or mobile phase don´t yield any peaks. Pure baseline.
-Autosampler works properly, verified by reproducible results with UV (substance for the test:cholesterylmyristate).
- All capillaries are steel, no peek involved.
What could be going on here?
The detector was purchased for this purpose and is also advertised for exactely this kind of application. I don´t want to do any derivatisation and neither do I have a GC.
I thank you very much for taking your time reading this and appreciate your input.
Best regards,
Klugeelse
I have enormous problems achieving reproducibilty - even over a time range of 8 hrs - with my HPLC-ELSD. Sample is trimyristine (saturated triglyceride with 3 C14-chains) in neat THF, concentration about 0,2mg/ml). The peak areas increase up to 30% within a day. I really don´t know what to do anymore and I have put in a decent amount of time to put it mildly.
HPLC-pump: Waters 515
autosampler: Waters 717 plus
detector: Alltech 3300 ELSD
Software: Clarity (Version?)
sample: trimyristine in THF, ca 0,2 mg/ml
column: ODS, 125mm*4mm; 5µm
eluentl: isocratic THF/ACN (45/55)
flow rate: 1,5 ml/min
injection volume: 10µl
run timet: 3 min
ELSD settings:
drift tube temperature: 40C
N2 flow: 1,6 l/min
gain: 1
I´ll give some more details that I think might be asked:
- the lab has an AC
- why 40C drift tube temperature? This is what the manufacturer advises for the given mobile phase, flow rate and sample concentration. Baseline looks awesome.
- The column is not heated, we don´t have an oven.
- All tests you can run are passed excellent according to the manual (i.e. gas flow, dirft tube temperature, clean optics.
- I run a hot cleaning (120C) for 1,5hrs; no changes.
- I took the detector apart and sonicated the nebulizer in MetOH/Water (50/50) . Didn´t change a thing. In the "light trap" I found some fluffy stuff that could be removed easily. Didn´t look like lipid at all.
- The peaks look good. Symmetrical, no fronting or tailing.
- Injection of neat THF or mobile phase don´t yield any peaks. Pure baseline.
-Autosampler works properly, verified by reproducible results with UV (substance for the test:cholesterylmyristate).
- All capillaries are steel, no peek involved.
What could be going on here?
The detector was purchased for this purpose and is also advertised for exactely this kind of application. I don´t want to do any derivatisation and neither do I have a GC.
I thank you very much for taking your time reading this and appreciate your input.
Best regards,
Klugeelse