IEX and ACN question

Posted: Mon Oct 18, 2010 3:57 pm
by pph31
Adding ACN can reduce the hydrophobic interaction in IEX. What's the mechanism? ACN interacts with the hydrophobic chain?

In typical RP-HPLC, ACN is used as mobile phase. Analytes interacts with the Carbon chain of the stationary phase. I never think about the ACN-reducing-the-hydrophobic-interaction story.

Is it contradictory?

Thanks

Posted: Tue Oct 19, 2010 9:05 am
by HW Mueller
The ACN is in competition with the hydrophobic analyte, in other words, the more ACN, the more the equilibrium of the analyte is shifted toward the mobile phase.