Hi AICMM,
The valve return to its original position after hydrogen peak eluted from molsieve column.
Carrier gas : N2
Detector: TCD
Constant Pressure and Isothermal temperature at 70C
Column: HaysepQ follow by Molsieve5A
Origin Position (Valve Off)
Sample injected (Valve On),
After 0.8mins backflush, flow reverse (Valve Off) to detector.
RT of hydrogen peak is around 0.6mins
The problem i encounter is the peak is shifting from left and right within repeated injection accompany with area increase when the peak shift to right, area drop when it shift to left.
The RT shift between 0.58 to 0.61.
We actually check for leak in system (column, fitting, valve) but could not find any.
Thank you