mxa, hfredricks,
For what it is worth, contact Cal-tech scientific in California. They may be able to help you with this problem. (Met them at a previous Pittcon.)
Best regards,
AICMM
Hey AICMM --
thanks for the tip. I got in contact with them, and they might help repair the item for less than Agilent would charge for the replacement.
I've figured out more stuff related to this problem, and ultimately I'll post a nice resolution/brain-dump, after I figure out a couple things.
For now, I've learned that the U+/U- cable connects from the analyzer board to the RF coil box. It's a three wire connector, and the middle wire is ground. According to the Agilent 1946 trouble shooting guide, you can test if the analyzer board is working by doing a "manual tune" in ChemStation and trying a couple different masses, while checking U+/U- with a volt-meter. (I believe measuring DC voltage). The connector carries high voltage, and I did shock myself yesterday, so be careful with it
It is easy to safely test the voltage with a multimeter though.
According to the troubleshooting guide, depending on the mass you "tune" to, you will see these U+ and U- values:
mass U+ U-
118 +30 -30
622 151 -151
922 223 -223
1522 369 -369
On my instrument, my U+ and U- are +796 and -266 (or +796 and -184), no matter what mass I set the manual tuner to. So they are wrong in magnitude (but correct in polarity). I'm assuming this is an analyzer board issue. The troubleshooting guide says that a broken analyzer board should be producing +0/-0 though. Hmmmm.... Maybe my board is broken in a different way.
Last question, in case anyone knows:
My analyzer board is an agilent G1946-60250, while the replacement that Agilent wants to provide is a G1946-65250. Anyone know the difference between these boards? I'm wondering if the latter one is improved somehow. I.e. does it make more sense for me to fix my 'old' board (~$1000), or throw down a little extra money and get an improved 'new' board (~$3300).
thanks for any hints.
-marc