Pressure leaps from run to run

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Hello everyboby.
This is my first time posting. I hope anyone can help me. My problem is i ran a serie of injections and the pressure raises from 1 injection to the next and drop back to normal again on the third. The pressure profiles look similar on all three except the second one is raised by 3 bar. A couples of injections later it happened again the pressure raised and turn back to normal on the next injection.

My conditions are

FLow: 1,000 ml/min
MP A: MeOH/ 28mM tertbutylammonium acetate 70/30(v/v)
MP B: MeOH
gradient
0 min 85% B
8 min 85% B
10 min 95% B
11 min 85% B
18 min 85% B

Anyone have an idea on what might cause this?

Thanks

What is your reequilibration time between 2 runs? What kind of RP column you are using?
Gerhard Kratz, Kratz_Gerhard@web.de

i am using a C-8, 150x4.6 mm with 3.5µm particles. 7mins at the end of the gradient is to reequillibrate the column. is it insufficient?

7 min for reequilibration at 1ml per min flow should be ok. Could be that your injection valve or your pump, the valves of A or B makes the troubble. If you use an autosampler please check if you have a leackage. Second, please check the flow of each valve A and B to see if both valves are working ok and each A and B has the exact flow. Please check also your flow rate while you get back pressure changes.
Gerhard Kratz, Kratz_Gerhard@web.de

I dont suspect the pumpvalves because the retention time for my two peaks are the same on all injections. I dont see any leaks at the injectionvalve. Right now im running another established method with different column and mobilephase. I will check the precision and see if the pressure leaps on this method aswell.

I´ve checked the flowrate from chanel A then B with a 10 ml volumetric flask. at 1 ml/min it took 10min and 10min 1 sec to fill the flask from chanel A recpective B. i injected a sample six times with the established method and the precision is fine (rsd=0,7%). No pressure leaps even thought i did 18 injections.

im suspecting the methanol, i dont filter it because its of HPLC grade. I saw some dust in the mobile phase B container. I´ll filter it from now on and see if the problem returns.
6 posts Page 1 of 1

Who is online

In total there are 347 users online :: 2 registered, 0 hidden and 345 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 345 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry