The method details are as follows:
Mobile phase: 1% Calcium chloride in 80/20 Methanol/Water
Extraction Solvent: 1% Calcium Chloride in Methanol
Detection: Ex318, Em390nm
1ml/min, 20µl, C18 column
RT of active = 5.0min
So far during the investigation blank solutions have generated positive and negative peaks at 2.7 and 5.0 minutes (i.e. interfering with the component of interest).
Methanol gave a positive peak at 2.7mn and negative at 5.0 (approx 2.5% of nominal)
Mobile phase gave negative at 2.7 positive at 5.0 (approx 0.2%)
Water (2 sources) gave negative at 2.7 and positive at 5.0 (approx 2.5%)
More confusing was that extraction solvent generated positive peaks at 2.7 and 5.0min (approx 8% of nominal) whilst a standard stock diluted into both mobile phase and extraction solvent gave similar responses.
Peaks in the extraction solvent (at approx 2.5%) have been observed on using alternative mobile phase and extraction solvent on another system (but with the same detector).
I am currently looking at different batches of Calcium Chloride but can anyone offer any other suggestions/possible causes.