Tropine is too volatile to be detected by ELSD under the separation condition. There are some posibilities that you can consider:
1. Select a proper column that can retain and separate tropine in TFA containing mobile phase. The adduct of tropine.TFA should be detected by ELSD or CAD. I need to point out that to obatin low ppm level sensitivity, your better shot is CAD not ELSD; or
2. Use MS or electro-chemical detection; or
3. Use IC and conductivity detection.
Thank you for your advice,it is very helpful for me! But I have no CAD ,only DAD, uv ,and RID! The tropine is inactive to UV , revealed end absorption by UV scan! I also tried testing by UV at 200、205nm……,but on result!
I know ELSD maybe have problem in sensitivity,but now because the instruments I really have no idea except with ELSD .However,UV is a good select if a end wavelength can be used.
Very pleased to know you! Can you give me more advice?