A coworker of mine is having an odd problem with her Agilent 5890 GC/FID. She noticed over the past few sequences that occasionaly an injection would occur that would result in a 200% recovery of her surrogate (in an extracted sample) or standard (calibration check), and subsequent re-injection would result in the expected amount. This problem has now become the norm - almost every injection is now almost perfectly double what her original calibration of the instrument indicates it should be. She changed the inlet liner and septum; the syringe; the injection tower; and finally the controller box. The problem is still occurring, and has not been resolved by any of the changes she made. She is running a splitless injection of diesel and motor oil standards and samples. She called Agilent, and they suspect her detector, but I would expect a problem in the detector to be more variable (e.g. not always double).
I know the post does not supply enough info, a cardinal sin of the Forum. However, I'm not well versed in GC, so I'm not sure what is pertinent in this post and what is not. So, if anyone could give me some ideas as to what might be causing this, or what information is needed to make more informed guesses, I would appreciate it.