how to add the second GC into the chemistation
Discussions about GC and other "gas phase" separation techniques.
14 posts
Page 1 of 1
Hey guys, we have one chemistation and want to control two GCs. Now we have this problem: Only one GC(instrument 1) shows up. Another GC(instrument 2) cannot show up. No shortcut at all. We use 5890 SERIES II GC-FID. does anyone have any suggestions? I will be most grateful! Thanks.
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
You will need a licence number for the second GC, then you can add it. I suggest that you contact Agilent on how to obtain a second licence number.
Gasman
Gasman
-
- Posts: 12
- Joined: Tue Jun 15, 2010 2:54 pm
thanks, but we added the license. The problem is that the second GC shortcut cannot show up. any thoughts?
You will need a licence number for the second GC, then you can add it. I suggest that you contact Agilent on how to obtain a second licence number.
Gasman
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
OK, you've added the second license. Did you go into Configuration Editor and add Instrument 2?
If you did that, and all you need is the desktop shortcut, find the .exe file using My Computer or Explorer, and just make a shortcut. Is the listing there under, start, programs, Chemstations?
If you did that, and all you need is the desktop shortcut, find the .exe file using My Computer or Explorer, and just make a shortcut. Is the listing there under, start, programs, Chemstations?
-
- Posts: 12
- Joined: Tue Jun 15, 2010 2:54 pm
the main problem is that we cannot find it in the listing under start, programs,chemstation even though we used configuration editor and added instrument 2. Thanks for any suggestions!!
OK, you've added the second license. Did you go into Configuration Editor and add Instrument 2?
If you did that, and all you need is the desktop shortcut, find the .exe file using My Computer or Explorer, and just make a shortcut. Is the listing there under, start, programs, Chemstations?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Right click on the Instrument 1 shortcut icon, go to properties. The path for the shortcut should be readily found. I'm not at work now, but write that down, then look for such using My Computer or Explorer.
It will be in HPCHEM, maybe under CORE, then under GC (if I remember correctly) and end with 1\a.exe or 1.exe for online and offline(or something similar, can't remember) for Instrument 1. So should be similar, like 2\a.exe or 2.exe (or something similar) for Instrument 2. If you can post a screen capture of your Instrument 2 under Configuration, that might help me to help you.
I've done this before, and it's straightforward. I don't believe you'll need to dig out the ChemStation CD (but could be wrong).
It will be in HPCHEM, maybe under CORE, then under GC (if I remember correctly) and end with 1\a.exe or 1.exe for online and offline(or something similar, can't remember) for Instrument 1. So should be similar, like 2\a.exe or 2.exe (or something similar) for Instrument 2. If you can post a screen capture of your Instrument 2 under Configuration, that might help me to help you.
I've done this before, and it's straightforward. I don't believe you'll need to dig out the ChemStation CD (but could be wrong).
-
- Posts: 19
- Joined: Wed Aug 19, 2009 3:02 pm
Not sure if this will work in your Chemstation version.
Insert the Chemstation CD and run setup.exe. The setup program will ask if you want to update the software. Go for it. Then, add a new instrument and configure it properly.
Insert the Chemstation CD and run setup.exe. The setup program will ask if you want to update the software. Go for it. Then, add a new instrument and configure it properly.
-
- Posts: 38
- Joined: Mon May 28, 2007 10:37 pm
I have had the same problem running Chemstation A.10.01 in WinXP. If you have your desktop icon for GC1 you can copy and paste it and in the new copy everything that's 1 change to 2. Then go to win.ini and copy the GC1 info and make a GC2 copy with the new path information. Not sure if there's anything else to change but this worked for me and is stable.
-
- Posts: 12
- Joined: Tue Jun 15, 2010 2:54 pm
We did the same thing today before i see your post:) But thank you!!!
We copy GC1 and rename as GC2. However, the new problem comes up. When i edit the method, i cannot change the injector B temperature. The error info is "Injector B temperature cannot be changed since Oven Tracking is on". I cannot find the Oven Tracking Box in the method. So i tried to use GC keyboard, however, each time i click "yellow button and Oven Track button", this info appears: INJ B NOT PROG INLET
Any thoughts? We use GC5890 SERIES II GC-FID.
Thanks for all who have given me very helpful suggestions. I really appreciate your time!!
We copy GC1 and rename as GC2. However, the new problem comes up. When i edit the method, i cannot change the injector B temperature. The error info is "Injector B temperature cannot be changed since Oven Tracking is on". I cannot find the Oven Tracking Box in the method. So i tried to use GC keyboard, however, each time i click "yellow button and Oven Track button", this info appears: INJ B NOT PROG INLET
Any thoughts? We use GC5890 SERIES II GC-FID.
Thanks for all who have given me very helpful suggestions. I really appreciate your time!!
I have had the same problem running Chemstation A.10.01 in WinXP. If you have your desktop icon for GC1 you can copy and paste it and in the new copy everything that's 1 change to 2. Then go to win.ini and copy the GC1 info and make a GC2 copy with the new path information. Not sure if there's anything else to change but this worked for me and is stable.
-
- Posts: 38
- Joined: Mon May 28, 2007 10:37 pm
Tiger, Did you solve the problem with the injector? I don't have a solution but if you find one please tell us how you fixed it.
-
- Posts: 12
- Joined: Mon Sep 20, 2010 7:36 pm
Do you think you control the GC is the GC you want to control? MaybeWe did the same thing today before i see your post:) But thank you!!!
We copy GC1 and rename as GC2. However, the new problem comes up. When i edit the method, i cannot change the injector B temperature. The error info is "Injector B temperature cannot be changed since Oven Tracking is on". I cannot find the Oven Tracking Box in the method. So i tried to use GC keyboard, however, each time i click "yellow button and Oven Track button", this info appears: INJ B NOT PROG INLET
Any thoughts? We use GC5890 SERIES II GC-FID.
Thanks for all who have given me very helpful suggestions. I really appreciate your time!!I have had the same problem running Chemstation A.10.01 in WinXP. If you have your desktop icon for GC1 you can copy and paste it and in the new copy everything that's 1 change to 2. Then go to win.ini and copy the GC1 info and make a GC2 copy with the new path information. Not sure if there's anything else to change but this worked for me and is stable.
You think you control the GC1 but actually you control the GC2 (and the GC2 do not have an on-column inlet.)
-
- Posts: 12
- Joined: Tue Jun 15, 2010 2:54 pm
i just manually set up the injector temp by using GC keyboard. This works for me since the injector temp is constant based on the method i use.
Tiger, Did you solve the problem with the injector? I don't have a solution but if you find one please tell us how you fixed it.
-
- Posts: 12
- Joined: Tue Jun 15, 2010 2:54 pm
No. We can control both GC and separately. No interference. You can run both GCs at the same time or you can run just one GC.
Do you think you control the GC is the GC you want to control? MaybeWe did the same thing today before i see your post:) But thank you!!!
We copy GC1 and rename as GC2. However, the new problem comes up. When i edit the method, i cannot change the injector B temperature. The error info is "Injector B temperature cannot be changed since Oven Tracking is on". I cannot find the Oven Tracking Box in the method. So i tried to use GC keyboard, however, each time i click "yellow button and Oven Track button", this info appears: INJ B NOT PROG INLET
Any thoughts? We use GC5890 SERIES II GC-FID.
Thanks for all who have given me very helpful suggestions. I really appreciate your time!!I have had the same problem running Chemstation A.10.01 in WinXP. If you have your desktop icon for GC1 you can copy and paste it and in the new copy everything that's 1 change to 2. Then go to win.ini and copy the GC1 info and make a GC2 copy with the new path information. Not sure if there's anything else to change but this worked for me and is stable.
You think you control the GC1 but actually you control the GC2 (and the GC2 do not have an on-column inlet.)
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
Tiger,
Have you tried loading the DEFCG.M method? Start with that and try building your method from there. Hope that helps.
Best regards.
Have you tried loading the DEFCG.M method? Start with that and try building your method from there. Hope that helps.
Best regards.
14 posts
Page 1 of 1
Who is online
In total there are 83 users online :: 1 registered, 0 hidden and 82 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 82 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 82 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science