Hi all,

We have to analyze chitosan oligomers. The conditions in the application is :

Column: aminopropyl ( 300x3.9x5), with ACN/H2H 70:30. RI detection.

But we do not have aminopropyl column but Amirex (HPX-87C) from BioRad.
Can I use instead and if yes, what are the condittions?

Thanks for any help

You can try the Aminex but I'm going to guess no way. Glucosamine comes out near the void and your larger oligomers are going to come out sooner so I would expect all oligomers of glucosamine to coelute.

The aminopropyl sounds promising.

You can try the Aminex but I'm going to guess no way. Glucosamine comes out near the void and your larger oligomers are going to come out sooner so I would expect all oligomers of glucosamine to coelute.

The aminopropyl sounds promising.
Thank you for the help. I will come back to the amiino column.
3 posts Page 1 of 1

Who is online

In total there are 40 users online :: 2 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry