Rganga » Mon Oct 04, 2010 12:35 pm
Thank you very much for your assistance,
Yes I decided to vent the system and give a clean to lenses, ion and detector, but I have never done this maintenance, I don’t feel secure to do it alone, so I asked for an engineer to come and at least I will do it with him this time and learn in order to do it by myself in the future.
The value of TIC are still by far very low comparing to the recommended values. And the detector diagnostic still failing at High Mass Noise step.
We never had results from this equipment, it is new but it was off for long time we did only some initial trials by injecting some fatty acids methyl esther. But we bought it mainly to do PCB and PBDEs analysis and if we are lucky we plan to try some dioxins. I was on training learning these techniques and I am just back to start setting up the techniques. But I found this problems.
We did not changed the column or anything else, as it was installed the first time.
The ms source setting is 200ºC, lens1: -25.1, lense3: -25.1, electron lens:15. Trap ofsset: -7. Electron energy: -71.2, emission current: 250-260 (setpoint at 250).
Thanks again, I am really desperado; I am waiting for the engineer and want to start with my samples soon.
Thanks