lmh » Tue Oct 05, 2010 3:10 pm
(1) pumping without a column, and no back pressure. You are right this is a problem, if you are using a low-pressure mixing quaternary pump. Binary high-pressure mixing pumps don't suffer. If you are using a quaternary pump with a gradient proportioning valve before the pump, you should buy a back-pressure regulator and fit it in place of the column. Alternatively you can use a long piece of thin tubing. If you do neither, it is likely your pump is supplying a fairly random mixture of any of the solvents... (you can test this by doing a gradient test with no back-pressure if you like; it will probably fail dismally).
(2) Alp's right; if you are looking to see if your instrument is performing in spec, you have to follow Waters' SOP as closely as possible.
(3) In theory, the solvent you are ionising the reserpine in is the solvent it was dissolved in, as there is no column, but in practise it will mix with the running solvent, which is why (1) and (2) are important.
(4) You don't need a strong proton donor (acid) for electrospray to work. Sugars will give a nice positive signal in water, even though they are insignificantly ionised in water, and water is only a weak source of protons. The electrospray process is electrochemical, and will form ions if ions are possible in solution.
(5) doubling the signal to noise is probably a piece of history related to pharmacopoia definitions: some people look at noise as the distance from bottom of baseline to top of baseline, while peak height is measured from the middle, so the logical view is that the bottom-to-top noise is twice what it ought to be if you are comparing with a peak. It really doesn't matter how you define things, but to know if your instrument is OK, you need to stick to the definition Waters used.
(6) sometimes reserpine dissoves a bit better if you sonicate it.