Dionex ICS-3000 Pump trouble shooting, Thanks

Posted: Mon Oct 04, 2010 3:02 pm
by Jady Wang
ICS-3000 pump would not run.
When I set up prime or flow rate, the pump stops.
It shows: Drive Current too Large. Camshaft 160 degree
Excessive drive current. Camshaft 160 degree

Posted: Wed Oct 06, 2010 2:18 am
by tom jupille
Disconnect system from the mains and then reconnect and re-start.

If the problem persists, it is not a do-it-yourself fix. I'm not familiar with that specific pump, but it sounds like you have either a broken or frozen piston. Completely blocked check valves might give the same message if the pressure transducer is downstream from the pump head (if the transducer is in the head, you would probably get a high-pressure warning first). And, of course, there is also the possibility of an electronic failure.

If this were my system, I'd call in the service engineer. :cry:

Posted: Thu Oct 07, 2010 9:43 pm
by Mark Albertson
Jady,

You have a possible blockage in the transfer line between the primary pump head and the secondary equilibration pump. As Tom has pointed out this is a service issue that cannot be rectified by the user and you should contact your local Dionex service rep.

Posted: Mon Oct 11, 2010 3:22 pm
by Jady Wang
Thanks Tom and Mark. I tried to disconnect, clean, combine last week. The problem is still there.

Posted: Mon Oct 11, 2010 6:00 pm
by Jady Wang
thanks again. I just fixed the pump.

Posted: Mon Oct 11, 2010 10:14 pm
by Mark Albertson
Hi Jady,

Was it a blockage in the transfer line between pump heads or something else??

Re:

Posted: Fri Feb 17, 2012 2:57 am
by mhjiang
thanks again. I just fixed the pump.
hi jady,can you tell me how you fixed the pump? The pump of dionex HPLC also has the problem of yours~ thanks a lot!

Re:

Posted: Fri Feb 17, 2012 3:01 am
by mhjiang
Hi Jady,

Was it a blockage in the transfer line between pump heads or something else??
hi MARK,can you tell me how i can do with my pump of HPLC? The pump of dionex P680 HPLC also has the problem of Jady ~ thanks a lot!

Re: Dionex ICS-3000 Pump trouble shooting, Thanks

Posted: Fri Feb 17, 2012 11:27 am
by Mark Albertson
Hi mhjiang,

There could be a blockage in the connector tube between the primary pump heads (the one with the check valves) and the secondary pump head. If a blockage occurs here pre-pressure transducer the pump sees the additional load as an over current. Other problems; if the pump was not maintained properly it could be worn bearings which will necessitate a service call.

Cheers

Mark

Re:

Posted: Mon Feb 20, 2012 11:46 pm
by williasj
thanks again. I just fixed the pump.
Hi Jady,
I am having the same problem, but I am sure that it is not a blockage (also occurs with the pump heads removed). Can you tell me how your solved your problem?
Thanks!

Re: Dionex ICS-3000 Pump trouble shooting, Thanks

Posted: Wed Feb 22, 2012 1:15 am
by Mark Albertson
Hi williasj,

If the problem is occurring with the pump heads removed its definitely not a blockage, I recommend you seek assistance from a Dionex Service Rep

Regards

Mark

Re: Dionex ICS-3000 Pump trouble shooting, Thanks

Posted: Thu Sep 13, 2012 4:30 pm
by yokes
Late to the party but...

I had this happen recently. A rollerbearing in the drive unit was failing and reground the cam on the pump drive.

Required a pump drive replacement.

Not a user replacement part.