by
krickos » Wed Oct 06, 2010 10:44 am
Might have misunderstood the question.
.....I can only use the USP standard ($620/50 mg) for the validation which is not preferable due to cost.
As long as you use the monograph I can not see that you get away from using the USP standard unless you qualify your own primary or secondary reference standard against the USP standard.
The Ph Eur standards are cheaper (79euros/50mg) and is supposed to come with a reference chromatogram in this case, the Ph Eur monographs contains more information about know impurities, however the procedure is not similar to the USP procedure so unsure on how informations you can extract from Ph Eur. But it could be intresting to compare both procedure and see if one is more suiteble than the other, also injecting both references on both procedures can be intresting as the references contains mixes of impurities.