Hello, I am a novice in experience with Hplc, I have ecountered something that has me stumped. When I inject my sample no peaks are detected. However, if i switch the valve to the other flow, the injection is detected during the realtime. It seems like the flow lines are crossed somehow, but I never physically changed anything. Has anyone had any experience with something like this before. Thanks

My system is:

Shimadzu

2 LC 20-AD pumps

Sil- 20 HT Autosampler

Spd-20A Detector

I run isocratic flow with one pump.

What do you mean when you say "switch the valve to the other flow?"

With a 2 pump Shimadzu, you probably have 4 lines that feed the two pumps. Make sure that the proper line is selected for the proper pump (or toss all 4 lines into your mobile phase bottle...)
Thanks,
DR
Image

-try to check your LC realtime in PUMP category maybe you used the other pump
-try also to check each line from the mobile phase reservoir to your pump.
- try to use gradient flow and test each individual line.

What is the backpressure before and after you've switched "the valve to the other flow"?

Best Regards
Learn Innovate and Share

Dancho Dikov
5 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry