Hope somebody here will have an answer for me.
It's been one month now that I can't analyse anything because of the presence of a peak in each of my analysis.
It started the day I changed the septum, as a very common maintenance, I did it the same way as usual etc...
Then on the next blanks I passed, there was this intense DEP peak and some small peaks of DOP and dimethyl phtalate.
Since then, the engineer in charge of my GCMS and I have been trying to : change the septum, the insert, the whool, cleaned and replaced the traps, change the column, change the gas bottle, the syringe... We ran a lots of blank analysis each time, with or without injection of solvent (I usually use ethanol), and with acetone as well. We tried to run only the MS and only the GC.
It seems to come from the GC not the MS. AFter all these trials, I don't know what's left to do... Do we have to change anything else in hte injection port?
The DEP peak keeps the same intensity and never decreases. Except when we programmed a special method high temperature to volatilise the DEP which is supposed to vaporize at 299deg. But afterwards, it never disappeared totally and it was back the same as before.
If you ever meet this problem before and/or have any suggestion, you are most welcome to express them here !!
Thank you in advance!
Delphine