anion exchange problem
Posted: Mon Jan 31, 2005 7:38 pm
I'm looking at recovery of a production step (diafiltration to remove CsCl). I'm using Resource Q which is a strong anion-exchange column. Gradient is from 300mM NaCl to 1500mM NaCl in 50mM Hepes pH 7.5.
My problem is I'm getting too high recovery of about 150%.
A) samples before diafiltration are in 5 mM Hepes pH 7.8 and high in CsCl. I have diluted these to 50mM Hepes pH 7.5 to reduce the salt content.
B) samples after diafiltration are in 5mM Hepes-20% glycerol pH 7.8 and without CsCl. These have been diluted to starting buffer: 300mM NaCl-50mM Hepes pH 7.5 and injected.
I thought that this was Cl- interference in A) samples and did some spikes to confirm correct dilution factor. Strangely, the spike recovery looks fine, around 100%?
Any ideas, what this could be?
What is the role of analyte counter ion? Can Cs+ for A) samples act differently as a counter ion, compared to Na+ for B) samples? What about glycerol?
Many thanks
My problem is I'm getting too high recovery of about 150%.
A) samples before diafiltration are in 5 mM Hepes pH 7.8 and high in CsCl. I have diluted these to 50mM Hepes pH 7.5 to reduce the salt content.
B) samples after diafiltration are in 5mM Hepes-20% glycerol pH 7.8 and without CsCl. These have been diluted to starting buffer: 300mM NaCl-50mM Hepes pH 7.5 and injected.
I thought that this was Cl- interference in A) samples and did some spikes to confirm correct dilution factor. Strangely, the spike recovery looks fine, around 100%?
Any ideas, what this could be?
What is the role of analyte counter ion? Can Cs+ for A) samples act differently as a counter ion, compared to Na+ for B) samples? What about glycerol?
Many thanks