Getting rid of extra time in method editor timetable

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
Greetings all! I have what I hope should amount to an easy topic. I am very familiar with the old chemstation software controlling the Agilent 1100 series LCs. However we recently upgraded to a 1260 with the new Agilent OpenLab CDS Chemstation Edition rev. C.01.02 [14]. I am having an issue writing a column equilibration method. The issue is I am trying to start the quat pump at 0.1 ml/min and slowly ramp up to 0.6ml/min. The problem I run into is there is always some extra time attached to the end of the timetable graph that reverts back to my original flowrate (0.1ml/min). This is not present in the timetable itself. The posttime is set to "Off".

Has anyone run into a similar situation with the new software and/or found a way around this issue?
is the method set only to ramp up?
have you tried to have 1-2 minutes more at 0.6 flow?
you can also do a "second run", with an isocratic method with flow at 0.6ml/min
Thank you for the reply,

Yes the method is a warm-up method to bring an Aminex HPX-87H sugar column up to operating conditions (both flow and temperature). Additional flow time at 0.6ml/min does not help as the timetable always automatically adds a few minutes as the original flow rate which results in an immediate drop in flow from 0.6 to 0.2 but then an immediate spike from 0.2 to 0.6 a few minutes later when the actual analytical method kicks in, and this spike in flow is obviously undesirable from a pressure and column life standpoint.
In the 1100 pump settings, you can set the flow ramp speed (I think what it's called - it's in the pump parameters section with the compressibility settings on the A.10 Chemstation) to (I think) 0.1mL/min/min. When you turn on the pump, it won't go immediately to 0.6mL/min - it will ramp at 0.1mL/min^2 to 0.6mL/min, then hold there. I know the B.04 Chemstation had the same feature available for the binary 1290 pump - it was in the Advanced Tab of the pump parameter settings and was called "Maximum flow gradient" and had a ramp up and ramp down setting that did the same thing. However, I've never seen the OpenLab CDS interface - is OpenLab so different from the B.04 Chemstation that this is no longer available, or is it just not available on the 1200-series quaternary pumps?

If it's still there, you can make a method called warm_up that you load without turning the pump on, that has the lowest ramp-up setting you can program. When you activate the method, the pump turns on and starts slowly making its way to 0.6mL/min. You can figure out how long you want that to go for, and the next line in the sequence (or the next method you activate) would be your normal method.
Time flies like an arrow. Fruit flies like a banana.
4 posts Page 1 of 1

Who is online

In total there are 177 users online :: 0 registered, 0 hidden and 177 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 177 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry