Column
Posted: Tue Sep 28, 2010 1:47 pm
by ossa97
hi...
I'm using a BDS HYPERSIL C18 5microns column for my method,,,,the column dies after around 50 injections:: these are my mobile phase conditions:: 90% Sodium Phosphate Buffet pH = 7.5, 10% ACN, with 0.1% TEA,Instrument Parameters: temp:35C, flow rate: 1ml/min, injection volume;10microl... i am just wondering what's killing the column, none of my conditions are too extreme,, please helppppp!!!!!
Posted: Tue Sep 28, 2010 2:15 pm
by Gerhard Kratz
I guess that Sodium phosphat buffer with 90% is killing your column if you use a brand new column and connect it. Shipping solvent is usually 70 % ACN. Do you see that back pressure is increasing? Please rinse your column with 90% water and 10% ACN before using your mobile phase, I mean your next new column. Please start with a lower flow rate and check back pressure. Good luck. BR Gerhard
Posted: Tue Sep 28, 2010 6:02 pm
by tom jupille
Are you sure the problem is mobile phase and not sample?
Because the column is deteriorating quickly, you can apply the following diagnostic test:
http://www.lcresources.com/resources/TSWiz/hs390.htm
Posted: Tue Sep 28, 2010 6:54 pm
by ossa97
The samples are all filtered on a 0.45 microm PVDF filter prior to injection...
Posted: Tue Sep 28, 2010 7:12 pm
by Uwe Neue
Do you add the TEA after you measured the buffer pH?
Posted: Tue Sep 28, 2010 7:22 pm
by LC_labrat
What is the concentration of the sodium phosphate? For example a 10mM solution running a 90%. Also if you premix the mobile phase does it precipitate out of solution?
Posted: Tue Sep 28, 2010 8:38 pm
by ossa97
I add the TEA before I measure the buffer pH.... and i adjust the pH with Sodium Hydroxide...
I use 0.02 M Sodium Phosphate...
Posted: Wed Sep 29, 2010 4:40 am
by Uwe Neue
Still tough to tell what the problem is. Some columns, especially those of an older style, are suffering under your mobile phase conditions. It is also possible that you have some gunk in your sample that we have not yet talked about...
about column details
Posted: Wed Sep 29, 2010 10:36 am
by cloudyarun
this is Arun from Aravind. now i am using C18 mm and 100A silca column..to wash the column we using ACN 100%.