actual peak collection before the PDA peak --??
Posted: Mon Jan 31, 2005 4:37 pm
I was checking the lag time between the fraction collector and Water's Empower software peak to set the wait time for the fraction collector. I noticed the seemingly impossible. I found the bulk of the injected standard at minute 13 (with hand collection or with a fraction collector), and the PDA software peak at minute 14.
The equipment being used is a Water's PDA 996 and a Water's 600 pump and controller.
This "pre-collection" occurs under only specific conditions and on one of 2 HPLCs: Silica column, 4%IPA-96%hexane, 1.7ml/min. isocratic. Other conditions tried with this std. and another compound actually gave a lag between the PDA software peak and the collection peak as expected.
When I pre-mix the 4-96 mobile phase myself, the situation doesn't occur.
On another lab computer with the same equipment, the situation doesn't occur. Also, switching columns between the two HPLC set-ups gives a retention time unique to the HPLC, not the column. In other words, Column A on HPLC A give a RT of 14 min. for my cmpd. Column B gives a RT of 11.2 min on HPLC B. However, Column A on HPLC B gives a RT of 11.2 min. And Column B on HPLC A gives a RT of 14 min.
The flow rate on both systems is identical and constant. The Water's technician checked the mixing on the HPLC and found no trouble.
Does anyone have any ideas? I am trying to develop a method involving peak collection which can be used on more than 1 HPLC system.
Thanks, E
Thanks, E