Dear all,

The senior technician in my lab suggest us to prime/purge the HPLC system with about 50 mL of hot water (70 degree C), to remove any possible salt precipitate in the tubing, prior to the start of analysis. We have followed this recommendation for about 6 months, our HPLC are working fine.

Last month, when the engineer from Waters visited our lab, they said the hot water might damage the HPLC system, without letting us know the exact reason.

I do not think it make sense, since the tubing, pump head, seal are quite durable material, they are supposed to withstand the hot water.

What is your opinion?

Regards,

Terry
<evil mode on>
Cleaning the instrument with hot water makes necessary repairs less frequent, therefore it is not wanted by the manufacturer...
<evil mode off>

I don't know any reason why hot water should damage your system. However, it should be necessary only if unused for some time and when you forgot to flush the system with salt-free solvents.

Jörg
Dera Jörg,

Thanks for your reply.

We have specified in our SOP to perform the hot water cleaning every time before use, as insisted by the 20-year experience technician. So we will stick to it, unless it is demonstrated that frequent hot water cleaning is not appropriate.

Regards,

Terry
Hi,

Hot water won't damage the system. It should be recommended any time to wash the system of salts. Good to hear you have a diligent seniot Technician.
I think the service Engineer may have thought you wash the system with Water and then leave it. But you do this before each use some no chance of any bacterial growth in the system.

Regards,
Bull76.
Your senior tech is correct. I have used hot water for many years, just remove the column first.
Then run dummy injections, which would clean out the low pressure paths too. Go back to your HPLC grade water , run the same method then go back to your initial conditions solvents, run that, then when the baseline looks stable, connect your column.
Good luck, and pass this on to others,
C.Tony Vella Royal British Legion
WWW.HPLCworks.net
858.663 751
Arte et Marte
Dear all,

Thanks for your clarification.

Regards,

Terry
6 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry