noisy baseline on purge & trap GC-MS
Posted: Mon Sep 27, 2010 8:01 pm
I’m new to this forum and spent some time looking through archives for a similar problem, but nothing solved my issue. I am looking for advice to fix an elevated baseline/noise problem. Can I clean this up or is this as good as it will get? Here is a lot of information. I came prepared.
I have attached 2 chromatograms(50ppb standard and method blank).
I have a very large CO2 peak at the beginning of my run and a baseline that I believe is causing me problems with early eluting compounds. I inherited this method and have made some alterations. I changed from industrial grade helium to ultra high purity grade and installed a big universal trap leading into the GC and purge and trap. I have installed a new repeller, insulator, liner-septum-gold seal, split vent trap, new column, new Supelco VOCARB 3000 K trap for P&T and cleaned the source with the same results. Also new 85/15 vespel/graphite ferrule. I am having issues determining the cause of this elevated baseline and running out of ideas. I also played with the P&T settings by adding dry purge time. Adding cryofocusing is not an option at this time. I ordered a Supelco carbopack B/carbosieve S-III trap to try as well to see if changing to carbosieve from carboxen would help.
I analyze wastewater samples via 8260B for 74 compounds ranging from Dichlorodifluoromethane to 1,2,4-trichlorobenzene with the usual suspects in between. I run an eight point calibration ranging from 20ppb to 400ppb. I am starting to have some larger problems with the early eluting compounds. Specifically Dichlorodifluoromethane and chloromethane. I think I will solve the chloromethane issue by lowering my upper calibration standard to 200ppb to correct the methanol suppression.
Here is my set-up:
Agilent6890N-GC
Agilent5973-MSD
Restek Rtx-624 (60m, 0.25mm diameter, 1.4um thickness) column
Thermal program
MSD Transfer Line: 280C
Oven: Max temp:250C Equilibration time:0.5min
4 minute solvent delay
Initial temp: 35C for 2.00min
15C/min to 210C, hold 12mins
25.67min run time
Front Inlet: Split mode-with Restek split liner no wool
Initial temp:200C
Pressure: 16.32 psi
Split ratio: 50:1
Split flow: 24.5mL/min
Gas saver: 20mL/min for 2mins
Column flow is 1.0mL/min
Tekmar 2016 purge and trap autosampler
Tekmar 3100 sample concentrator
Supelco VOCARB 3000 K trap
Purge and Trap settings
Line Temp:100C
Valve Temp: 100C
MCS Line Temp: 40C
Purge ready temp: 36C
Purge temp: 36C
PrePurge time: 3mins
Sample Preheat time: 5mins
Sample Preheat Temp: 40C
Purge time: 11mins
Dry purge time: 0mins
GC Start option:Start of Desorb
GC cycle time: 0mins
Desorb Preheat Temp: 220C
Desorb time: 4mins
Desorb Temp: 225C
Sample drain: on
Bake time: 10mins
Bake temp: 225C
Bake Gas bypass: on
Bake Gas bypass delay time: 2mins
MCS bake temp: 300C
20XX Valve temp: 100C
20XX Line temp:100C
Please Help.
Thanks,
Kevin





I have attached 2 chromatograms(50ppb standard and method blank).
I have a very large CO2 peak at the beginning of my run and a baseline that I believe is causing me problems with early eluting compounds. I inherited this method and have made some alterations. I changed from industrial grade helium to ultra high purity grade and installed a big universal trap leading into the GC and purge and trap. I have installed a new repeller, insulator, liner-septum-gold seal, split vent trap, new column, new Supelco VOCARB 3000 K trap for P&T and cleaned the source with the same results. Also new 85/15 vespel/graphite ferrule. I am having issues determining the cause of this elevated baseline and running out of ideas. I also played with the P&T settings by adding dry purge time. Adding cryofocusing is not an option at this time. I ordered a Supelco carbopack B/carbosieve S-III trap to try as well to see if changing to carbosieve from carboxen would help.
I analyze wastewater samples via 8260B for 74 compounds ranging from Dichlorodifluoromethane to 1,2,4-trichlorobenzene with the usual suspects in between. I run an eight point calibration ranging from 20ppb to 400ppb. I am starting to have some larger problems with the early eluting compounds. Specifically Dichlorodifluoromethane and chloromethane. I think I will solve the chloromethane issue by lowering my upper calibration standard to 200ppb to correct the methanol suppression.
Here is my set-up:
Agilent6890N-GC
Agilent5973-MSD
Restek Rtx-624 (60m, 0.25mm diameter, 1.4um thickness) column
Thermal program
MSD Transfer Line: 280C
Oven: Max temp:250C Equilibration time:0.5min
4 minute solvent delay
Initial temp: 35C for 2.00min
15C/min to 210C, hold 12mins
25.67min run time
Front Inlet: Split mode-with Restek split liner no wool
Initial temp:200C
Pressure: 16.32 psi
Split ratio: 50:1
Split flow: 24.5mL/min
Gas saver: 20mL/min for 2mins
Column flow is 1.0mL/min
Tekmar 2016 purge and trap autosampler
Tekmar 3100 sample concentrator
Supelco VOCARB 3000 K trap
Purge and Trap settings
Line Temp:100C
Valve Temp: 100C
MCS Line Temp: 40C
Purge ready temp: 36C
Purge temp: 36C
PrePurge time: 3mins
Sample Preheat time: 5mins
Sample Preheat Temp: 40C
Purge time: 11mins
Dry purge time: 0mins
GC Start option:Start of Desorb
GC cycle time: 0mins
Desorb Preheat Temp: 220C
Desorb time: 4mins
Desorb Temp: 225C
Sample drain: on
Bake time: 10mins
Bake temp: 225C
Bake Gas bypass: on
Bake Gas bypass delay time: 2mins
MCS bake temp: 300C
20XX Valve temp: 100C
20XX Line temp:100C
Please Help.
Thanks,
Kevin