Its Pka is 1.8.
m.P-A: WATER: NETHANOL (95:5)
M.P-B: METHANOL
COLUMN:Zorbax-XDB-C18-150X4.6mm,5mc
FLow ;1 mL/min
UV:254
Gradient
T/%B : 0/20, 10/30, 25/40,30/60,40/60,45/20
The method was successfully tranferred from supplier lab to ours.The peak rt is 15 min.When one inject sample ,no problem for the first injection.However as injections progress a peakat rrt 0.9 starts elluting.But no problems of stability either.The method runs good in originating lab.What could be the problem?