Tray malfunction question

Discussions about GC and other "gas phase" separation techniques.

7 posts Page 1 of 1
Has anyone ever had this happen to them right after the tray puts the sample vial in the turret?

Image

and then you click retry...

Image

I did a search on Google for "tray malfunction" and got nothing useful.

The instrument is an HP5890 and the tray is a 18596B.

Any ideas? It's really annoying seeing that first screen as soon as you walk into the lab in the morning.

Thanks,

John

Did it actually get the vial into the turret?

I have seen this several times where the arm actually dropped the vial or was not able to pick up the vial from the tray.

Did it actually get the vial into the turret?
Yes, it did. Strange. I get four flashes on the controller box also.

EDIT: I just found this:

Tray Errors
4 Single Flashes: Radial Axis Error


But if this was a real issue, it wouldn't work on the retry...right?

John

Did you see this?

http://www.chem.agilent.com/Library/Sup ... A16006.pdf
bottom of page 11

Hope this helps.

Did you see this?

http://www.chem.agilent.com/Library/Sup ... A16006.pdf
bottom of page 11

Hope this helps.
Yeah, I just saw that...but nothing is obstructing the arm, it's not grinding gears or anything.

I came in this morning to see that Tray Malfuntion dialog box (which happened 30 minutes after I left last night).

Started another run this morning, and it's gone through 4 injections so far without an error.

Seems to be inconsistent.

I'll try troubleshooting the tray when I get a chance. I have way too many samples to run to have one instrument down while I'm here, though.

Thanks, Will!

John

The only other thing I could suggest is the belt below the arm could be worn and slipping. That issue was very inconsistent and problematic, causing all sorts of issues with the arm. We also had multiple samples to run, it was very
very annoying. Such as it is though, the "beauty" of instrumentation is spending half of your time trying to figure out what in the world is going on.

Good luck.

chemwipe,

I can recommend someone who fixes trays and towers. Very reasonable.

aicmm at flash dot net

Best regards.
7 posts Page 1 of 1

Who is online

In total there are 83 users online :: 1 registered, 0 hidden and 82 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 82 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry