Controlling actuators with Varian 24 v external events.
Posted: Tue Sep 21, 2010 3:21 pm
by Steve P
I'd like to control a number of Valco electric and microelectric two position actuators from a Varian GC-450 using Galaxie software. The Valco actuators can be switched with a contact closure but the Varian external event boards in the GC have 24 volt power on/off relays. I'm guessing that I'll have to use intermediate relays powered from the 24 volt signal to provide the contact closure. I wonder if there is a standard solution to this? Would mechanical or solid state relays be most suitable?
Thanks in advance
Steve
Posted: Tue Sep 21, 2010 10:41 pm
by chromatographer1
Call Valco.
Really.
or Call Varian.
Really.
Who else knows their hardware better than they?
Free advice is often worth what you pay for it.
Really.
best wishes,
Posted: Wed Oct 06, 2010 2:25 pm
by srattier
If your Valco valve requires a contact closure, then the 450-GC relays are what you need. The 450-GC relay are 24VDC meaning they can open/close a voltage up to 24V... I would be really surprised if the Valco is using higher voltage on the contact input. I think you can directly connect your Valco actuator on your GC relays. If you have any doubt, you'd better ask to Bruker now (450-GC is now Bruker) and Valco for specification about the GC relays and the actuator.