Glycoalkaloids
Posted: Fri Sep 17, 2010 11:05 am
by MTM
I'm trying to apply a method for quantification of glycoalkaloids (Tomatine and tomatidine).
I'm using a HPLC/PAD unit (Gilson) and a ACE 3 C18-AR.
Detection: 202 and 205 nm
Which reference value I should use for this wavelenght?
Thanks for your atention.
Posted: Fri Sep 17, 2010 2:32 pm
by carls
What is your mobile phase composition? Is it retained with RPLC? Have you considered HILIC?
Posted: Fri Sep 17, 2010 10:15 pm
by MTM
I'm using Methanol and CH3COOH
Posted: Sat Sep 18, 2010 12:21 am
by carls
Please provide your chromatographic conditions. Is your method isocratic or gradient? What are the column dimensions, temperature and flow rate?
What are the retention times of your 2 analytes? Please post a chromtogram if possible.
Instructions for embedding chromatograms in a post are shown here:
viewtopic.php?t=2617
"Which reference value I should use for this wavelength?"
Do you mean what detector reference wavelength should be used? If so, why do you think you need one? A reference wavelength is not always needed or helpful. Try it first without a reference. If the baseline drift is too high, 360nm +/- 50nm commonly used
Posted: Mon Sep 20, 2010 10:11 am
by lmh
You don't need hilic for tomatine and tomatidine. They are well-retained on C18. I don't envy you trying to detect them by UV, though. I use LC-MS for the job.
Posted: Mon Sep 20, 2010 10:28 am
by Bryan Evans
Below is an LC-MS application for a-solanine & a-chaconine from potato. This separation was done using Cadenza CD-C18:
http://www.imtaktusa.com/site_media/fil ... TI189E.pdf
As already stated, it will be difficult to do this analysis with UV detection.