easy way for increasing resolutions
Posted: Thu Sep 16, 2010 12:38 pm
The theory
Instead of using a longer column, resolutions in GC can be increased substantially by providing backpressure on the column and slowing down the linear velocity to its new optimum. The longitudinal diffusion (B in Van Deemter equation) will be less and separations improve. This may be as well applicable to LC separations.
Backpressure in GC can be provided by mounting a thinner capillary -no stationary phase needed here- between column and detector.
In literature about GCxGC, where two columns are coupled in this manner, it is often stated that a peak gets 'focused'. Focussing means the same as decreasing longitudinal diffusion.
The experiment
Isopropylbenzene and n-propylbenzene were injected at different flows on a 30 m x 0.25 mm x 0.25 um DB-1 column connected to a BPX-50 column of i m x 0.1 mm x 0.1 um. Resolutions were calculated at 10% peak height. Linear velocities were calculated as the quotient of column length and retention time of methane. The optimum linear velocities (U opt.) were assigned to the highest resolutions. Pressures were calculated with Hagen-Poisseuille and the universal gas law (see http://www.chem.hope.edu/~polik/Chem345 ... osity.html)
i (m) 0; 1.5; 3.0
U opt. (cm/s) 30 (lit.); 15; 12.4
inlet p (bar) 1.95; 3.01; 4.05
back p (bar) 1.01; 2.52; 3.64
Resolution n.a.; 4.9; 6.1
You can see backpressure and resolution increase and the optimum velocity drop. The attached BPX50 column has little selectivity between iso- and n-propylbenzene, but it decreases the pressuredrop over the DB-1 column and hence increases the resolution. Unfortunately there is no resolution data without the BPX50 column attached.
The questions
Are you familiar with using backpressure for improving separations?
If so, how do you restrict your column outlet?
If not, why don't you try it?
Instead of using a longer column, resolutions in GC can be increased substantially by providing backpressure on the column and slowing down the linear velocity to its new optimum. The longitudinal diffusion (B in Van Deemter equation) will be less and separations improve. This may be as well applicable to LC separations.
Backpressure in GC can be provided by mounting a thinner capillary -no stationary phase needed here- between column and detector.
In literature about GCxGC, where two columns are coupled in this manner, it is often stated that a peak gets 'focused'. Focussing means the same as decreasing longitudinal diffusion.
The experiment
Isopropylbenzene and n-propylbenzene were injected at different flows on a 30 m x 0.25 mm x 0.25 um DB-1 column connected to a BPX-50 column of i m x 0.1 mm x 0.1 um. Resolutions were calculated at 10% peak height. Linear velocities were calculated as the quotient of column length and retention time of methane. The optimum linear velocities (U opt.) were assigned to the highest resolutions. Pressures were calculated with Hagen-Poisseuille and the universal gas law (see http://www.chem.hope.edu/~polik/Chem345 ... osity.html)
i (m) 0; 1.5; 3.0
U opt. (cm/s) 30 (lit.); 15; 12.4
inlet p (bar) 1.95; 3.01; 4.05
back p (bar) 1.01; 2.52; 3.64
Resolution n.a.; 4.9; 6.1
You can see backpressure and resolution increase and the optimum velocity drop. The attached BPX50 column has little selectivity between iso- and n-propylbenzene, but it decreases the pressuredrop over the DB-1 column and hence increases the resolution. Unfortunately there is no resolution data without the BPX50 column attached.
The questions
Are you familiar with using backpressure for improving separations?
If so, how do you restrict your column outlet?
If not, why don't you try it?