Thanks in advance for any thoughts you may share about this.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Yep, that was the point.if it passes all sensible validation, then the method is fine whatever aspect of the peak you quantified.
It turns out this is the case with this client test method. The offical document shows example chromatograms of standards and only two sample types, but this method is designed to be used for a variety of products that contain biotin. When I looked back at the validation report which had example chromatograms for all the different products, resolution issues became apparent.Maybe you saw only chroms of standard or nearly ideal samples.
I agree about solution stability. You could make a case for robustness needing revalidation, because integration may be affected by things like peak shape and critical resolution.My feeling is accuracy, precision and linearity will need to be reassessed. Do robustness and solution stability studies need to be reassessed also?