Baseline dip at high temperatures
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
I have been having a problem with my baseline starting high after the air peak then difting down even lower than my baseline. Then after the method is complete the baseline returns to normal. I only noticed this after switching to high sensitivity mode on the TCD. Any suggestions?
-
- Posts: 391
- Joined: Tue Dec 30, 2008 2:57 pm
What carrier gas, what make up gas, what analyte/sample matrix?
-
- Posts: 2
- Joined: Fri Sep 17, 2010 5:32 pm
We have two 5890s that we have manufactured our own headspace analyzer for. We use helium as our carrier gas and have bypassed the purge system. They have been setup like this for decades, but lately just one of them has started to exhibit this baseline dip. The samples we test are small round punchouts of coated material for retained solvents. Usually MEK, Toluene, alcohols etc.
3 posts
Page 1 of 1
Who is online
In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: CharlesPar and 15 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: CharlesPar and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science