- Posts: 3
- Joined: Thu Jun 14, 2012 11:48 pm
My name is Gilberto, I´m from Fortaleza, Brasil. We´ve here in our laboratory one Shimadzu SCL 10AVp System Controller, LC 10 AT Pump Control and SPD 10A UV Detector in one simple isocratic system, connected with one PC Computer with Windows XP SP3. We use CLASS VP 6.1 SP2. Recently this computer broke, and called me to fix it. So, we try connect this system between RS232C to computer but CLASS VP shows NOT CONNECTED, in a Instrument menu we make a correct choice to COM1, and computer with a same port enabled COM1 - 3F8/IRQ4, but doesn´t connect... We try change like as described in a Shimadzu manual about COM1 parameters, like as Velocity port in a 19200bps, parity: Even and... Stops bits in 1, but not sucess. When I try connect with LCMaintenance.exe included in a CLASS-VP software package, we hear one beep from SCL controller and this program shows pump and detector parameters, like as hours... without problem.
Anybody here have one help to us?
Comunication between SCL and Pump and UV are OK, with optical cable.
Thanks a lot.
Gilberto Mendes,