- Posts: 21
- Joined: Thu Jul 02, 2009 9:06 pm
What I do is in a electropolished SS 300ml vessel, I inject n-dodecane either pure or diluted in dichloromethane into the vessel with a 5 ul syringe. I then fill the vessel with high purity helium around 20 psi. The vessel is connected to the sample loop with treated or fused silica tubing, where all the tubing connected to the injection port is heated to 150 deg. C.
The fuel conc. is usually around 3-400ppm. Now the problem arises is that if I inject into the GC 3 ppm the calibration will be different if I make a new mixture and inject 400ppm into the GC. Basically the calibrations are not linear.
Is there a better way of doing this calibration. According to my experimental data, my calibrations for n-dodecane are off by 60-80%.
Would anyone know if there are limitations on the gas injection of a heavy fuel on the PLOT-Q or is the sample loop for the column? Could the column be damaged or too old?