New to forum and relatively new to GC, I thought I'd ask about issues with peaks in my blank. Let me start from the beginning.
I was working on transfering a method on DMF detection. The solvent the method calls for is Acetonitrile, the column: an Agilent RES-SOLV column. Anyway, my first trial injection of Acetonitrile was clean. Next one (my standard) showed a peak (yay!). After performing a sample set of my standards, my %rsd didn't pass and my blanks showed a "DMF" peak. After conditioning the column for a few hours and cleaning out the syringe, there's still a peak in my blank. I obtained another blank sample to inject; still a peak. I tried a different solvent manufacturer; still a peak. Okay, maybe there's just an impurity in Acetonitrile that elutes at the same retention time as DMF.
Unfortunately I'm pressed for time, so I was instructed to move on to another method that uses a different column and solvent. This one uses DMSO as the solvent and a DB-WaxPlus column. I've ran this method before to test a different product and it worked beautifully. Now everything's set up, cleaned out my syringe, my solutions are ready to go and... guess what... there's a peak in the blank!! Again?! So I changed my liner (which was clean but I changed it anyway), and ran a fresh solvent blank. Peak again.
Currently I'm conditioning the column; hopefully I won't have a peak again. My next step (I think) is to change the septum (which was changed about two weeks ago). But there must be something I'm clearly missing. Has anyone else ran into this problem before? If so, what steps did you take to clear it up?