I have to optimize an HPLC separation of Flavomycine Complex which mainly consists of 5 isomers of moenomycin:

The original method was taken from somewhere a long time ago - it uses Hypersil BDS-C18, 150 x 4.6 mm, 5um column and mobile phase consisting of 15 mM Na-heptane sulphonate as ion-pair reagent, and 50 mM phospate buffer with pH=7.0. Flow rate was 0.9 ml/min, Vinj=20ul, UV=256 nm. These conditions give the following chomatogram with 30 min run time:

Because this routine analysis takes a lot of time we want to optimize the method in terms of analysis time mainly, and Rs of min = 1.8 is required. We have to use conventional HPLC system.
Because I work for an Agilent distributor we initially chose Eclipse XDB C-18, 75 x 2.1 mm ID, 3.5 um. I didn't manage to achieve analysis time below 15 min with this column. Now I'm working with Poroshell 120 EC-18, 100 x 3.0 mm ID, 2.7 um, the best result received is below (with run time 10 min) :

0.7 mL/min mobile phase of 10 mM AmAcetate + 25 mM Na-heptane sulphonate (53%), AcN (35%) and MeOH (12%).
And then began problems with method stability - shifts in the retentions and, as a consequence loss of Rs. All temperatures are stabilized except that of solvents before entering the pump.
I'm open for all opinions and I continue to look for points influencing separation process reproducibility.
___________________________________
Alexander