unusual EIC

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Virtual Colleagues.

I when I extract the IC for 477 m/z ion from the mass spectra I obtain the unusual (to me) plot (last of 4). The first plot is the TIC, second plot is the MS of the elaidic acid peak (1st integrated peak). Third plot is the EIC for 283 m/z ion, which is elaidic acid. This plot is what I expect to obtain. Given the relative abundance of 283 and 477 ions are similar, I do not understand the difference is the EICs. Thanks for the feedback.

The data was collected on an Agilent LC-MSD SL 1956B. Mobile phase is 70/30/0.05, ACN/Water/TFA. Zorbax SB-CN column. Postive ES mode.

[/img]Image

Image[/img]

regards,

I'm surprised no one has replied.
477 is probably a background ion in your runs. It is present at high levels all the time, which is why you see it in the spectrum, and why the EIC looks like it does. There is a negative peak in the EIC of 477 at the time the major peak elutes from the TIC, probably indicating that whatever caused the major peak suppresses ionisation of the background ion.

Imh,

Thanks for your input.

regards
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry