ISO9377-2 oil index or oil in water: septum bleed

Posted: Wed Jan 26, 2005 11:50 pm
by remirama
Recently, we started to use ISO 9377-2 method for oil index in water (waste water) determination. System: Varian GC FID, 15mx0.25x0.25 VF1 column, on column injector. 5m of 0.53 precolumn. The problem is very high septum bleed - signal peaks in range of C10-C40 are in order of 30 mV (0.2 mg/l) and septum peaks are in the same order (going out ~150C. Temp. program: 50C for 2 min. and 60C/min to 320C hold for 2 min. Flow - 4 to 2 ml/min. I've tried changing almost everything, but nothing helps. And another big problem - C40/C20 ratio is very small = 0.35. It should be >0.8 according to standard. Maybe someone had the same problems? I will be glad to get any help.

Posted: Mon Feb 14, 2005 2:30 pm
by csolaz
Dear Remirama,

I work with a method based in 9377-2, but I work in splitless mode, at 340°C, column flow 7.4 ml/min, dtector at 325 °C. My initial temperature is 40°C for 1.50 min, then up to 60°C at 5°C/min and finally up to 315°C at 15°C/min. In such conditions C40/C20 ratio is nearly or higher than 0.80 and no bleeding or not significant.

I hope this will help you

Carmen

for Carmen

Posted: Tue Feb 15, 2005 8:52 pm
by remirama
Thank you for your answer, but just to be sure two more questions. Are you turning on split after some time (~5 min)? Is injector isothermal (all the time 340 or temperature program)?
Now i'm working with such configuration: injector 40 to 340C, split is on initialy (vents out hexane, very nice peak), after 10s split is off for 2 min. and after on for remaining time. Injection volume - 2ul. Temp. program: 30C for 2 min, after to 330C, at 30C/min and holds for 2 min. C40 comes at 13-th minute (15mx0.25x0.25 column). C40/C20 = 0.87. With longer period of initial split on, discrimination appears for the low boiling point hydrocarbons.
I think, it is not possible to use on column injector for that analysis, at least for injectors that use hot oncolumn and do not have septum purge.
Best regards,
remi

Posted: Tue Feb 15, 2005 9:54 pm
by varian
what varian do you have?

varian

Posted: Wed Feb 16, 2005 3:25 pm
by remirama
Varian CP3800, EFC type 3 (flow) and type 1(pressure), FID detector, 1079 split/splitless(PTV) and 1041 on-column injector (isothermal), 8410 autosampler.

Posted: Wed Feb 16, 2005 4:53 pm
by varian
Dear Remirama,
I have the same system Varian CP3800.
I do C10 to C60 on it.
split ratio 1/50 after .5 min
temp. prog. start 50(2) ramp 30C/min to 350C(0)
Injector temp. 350C
detector 360C
Flow 10ml/min
used high temp column db1 (30mx 0.32id x0.1 column)
it works great!!!!!!!!have fun

Re: for Carmen

Posted: Thu Feb 17, 2005 12:30 pm
by csolaz
dear Remirama,

I don't apply split ratio an my injector is isothermal.

Best regards

carmen


Thank you for your answer, but just to be sure two more questions. Are you turning on split after some time (~5 min)? Is injector isothermal (all the time 340 or temperature program)?
Now i'm working with such configuration: injector 40 to 340C, split is on initialy (vents out hexane, very nice peak), after 10s split is off for 2 min. and after on for remaining time. Injection volume - 2ul. Temp. program: 30C for 2 min, after to 330C, at 30C/min and holds for 2 min. C40 comes at 13-th minute (15mx0.25x0.25 column). C40/C20 = 0.87. With longer period of initial split on, discrimination appears for the low boiling point hydrocarbons.
I think, it is not possible to use on column injector for that analysis, at least for injectors that use hot oncolumn and do not have septum purge.
Best regards,
remi

Posted: Mon Mar 07, 2005 11:44 am
by renata
Dear Remirama,
try to look at DANI web site (www.danispa.it). There is an application note with a chromatogram and instrument parameter required. I use a DANI system and I'm satisfied.