ISO9377-2 oil index or oil in water: septum bleed
Posted: Wed Jan 26, 2005 11:50 pm
by remirama
Recently, we started to use ISO 9377-2 method for oil index in water (waste water) determination. System: Varian GC FID, 15mx0.25x0.25 VF1 column, on column injector. 5m of 0.53 precolumn. The problem is very high septum bleed - signal peaks in range of C10-C40 are in order of 30 mV (0.2 mg/l) and septum peaks are in the same order (going out ~150C. Temp. program: 50C for 2 min. and 60C/min to 320C hold for 2 min. Flow - 4 to 2 ml/min. I've tried changing almost everything, but nothing helps. And another big problem - C40/C20 ratio is very small = 0.35. It should be >0.8 according to standard. Maybe someone had the same problems? I will be glad to get any help.
Posted: Mon Feb 14, 2005 2:30 pm
by csolaz
Dear Remirama,
I work with a method based in 9377-2, but I work in splitless mode, at 340°C, column flow 7.4 ml/min, dtector at 325 °C. My initial temperature is 40°C for 1.50 min, then up to 60°C at 5°C/min and finally up to 315°C at 15°C/min. In such conditions C40/C20 ratio is nearly or higher than 0.80 and no bleeding or not significant.
I hope this will help you
Carmen
for Carmen
Posted: Tue Feb 15, 2005 8:52 pm
by remirama
Thank you for your answer, but just to be sure two more questions. Are you turning on split after some time (~5 min)? Is injector isothermal (all the time 340 or temperature program)?
Now i'm working with such configuration: injector 40 to 340C, split is on initialy (vents out hexane, very nice peak), after 10s split is off for 2 min. and after on for remaining time. Injection volume - 2ul. Temp. program: 30C for 2 min, after to 330C, at 30C/min and holds for 2 min. C40 comes at 13-th minute (15mx0.25x0.25 column). C40/C20 = 0.87. With longer period of initial split on, discrimination appears for the low boiling point hydrocarbons.
I think, it is not possible to use on column injector for that analysis, at least for injectors that use hot oncolumn and do not have septum purge.
Best regards,
remi
Posted: Tue Feb 15, 2005 9:54 pm
by varian
what varian do you have?
varian
Posted: Wed Feb 16, 2005 3:25 pm
by remirama
Varian CP3800, EFC type 3 (flow) and type 1(pressure), FID detector, 1079 split/splitless(PTV) and 1041 on-column injector (isothermal), 8410 autosampler.
Posted: Wed Feb 16, 2005 4:53 pm
by varian
Dear Remirama,
I have the same system Varian CP3800.
I do C10 to C60 on it.
split ratio 1/50 after .5 min
temp. prog. start 50(2) ramp 30C/min to 350C(0)
Injector temp. 350C
detector 360C
Flow 10ml/min
used high temp column db1 (30mx 0.32id x0.1 column)
it works great!!!!!!!!have fun
Re: for Carmen
Posted: Thu Feb 17, 2005 12:30 pm
by csolaz
dear Remirama,
I don't apply split ratio an my injector is isothermal.
Best regards
carmen
Best regards,
remi
Posted: Mon Mar 07, 2005 11:44 am
by renata
Dear Remirama,
try to look at DANI web site (
www.danispa.it). There is an application note with a chromatogram and instrument parameter required. I use a DANI system and I'm satisfied.