Any one try to a on-column injection from
a split/splitless inlet? in splitless mode with
a appropriate liner may It work ?

The test method that I'm trying is
ASTM D 6584, using a DB-5 HT
Best Regards
Daniel H. Sagarra
Field Application Engineer
Buenos Aires, Argentina
www.flowmeet.com

Restek have an inlet liner specifically for this - check out their web site.

Peter
Peter Apps

Just be carefull of trying to this modification. You will need to find a syringe with a needle long enough to inject your sample on the column inside the oven. You cannot do on-column injections using a high injector temp.

Don't you also need to modify the injector on such a system? For example, if using a standard septum injector, you require a needle robust (and thick enough) to pierce the septum but thin enough to enter the column-could be difficult. I guess this would work with wide bore columns though.
4 posts Page 1 of 1

Who is online

In total there are 96 users online :: 0 registered, 0 hidden and 96 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 96 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry