Looking for conductivity monitor/detector
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
Can anyone suggest where I can find a new conductivity monitor/detector like the LKB/Pharmacia/GE unit which has been discontinued with no replacement?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Dionex, Metrohm, Waters, Shimadzu come to mind. There are probably others as well.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 148
- Joined: Thu Nov 02, 2006 10:39 pm
We just bought a standalone GE/Amersham conductivity detector to add onto an Agilent stack. Are you talking about something different because the GE guy right next to me said they have no plans to quit selling them.
I used to have a Wescan conductivity meter.
I used to have a Wescan conductivity meter.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Dcree,
Check this out:
http://www.gelifesciences.com/aptrix/up ... one=Labsep
In addition to the conductivity detection/monitoring you get the pH as well. You might ask: Why should I monitor the pH? Well, if you’d like to safe time and material you might go for on-line buffer preparation. Another use for monitoring the pH could be in investigation/trouble-shooting situations. Anyway, you may choose to monitor the pH and you may choose not to, but the extra functionality is there.
Another possibility, especially if you only need a single UV wavelength, it might be adequate for you to have all 3 measurements capability (UV, conductivity and pH) integrated in the same instrument.
http://www.gelifesciences.com/aptrix/up ... uleid=8933
Best Regards
Check this out:
http://www.gelifesciences.com/aptrix/up ... one=Labsep
In addition to the conductivity detection/monitoring you get the pH as well. You might ask: Why should I monitor the pH? Well, if you’d like to safe time and material you might go for on-line buffer preparation. Another use for monitoring the pH could be in investigation/trouble-shooting situations. Anyway, you may choose to monitor the pH and you may choose not to, but the extra functionality is there.
Another possibility, especially if you only need a single UV wavelength, it might be adequate for you to have all 3 measurements capability (UV, conductivity and pH) integrated in the same instrument.
http://www.gelifesciences.com/aptrix/up ... uleid=8933
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 7
- Joined: Mon Jul 16, 2007 1:35 pm
The GE monitor I was looking for is at http://www.gelifesciences.com/aptrix/up ... one=Labsep. Customer Service told me it's discontinued without a replacement. The pH/C-900 looks like it might work.
Thanks to all for your help.
Thanks to all for your help.
5 posts
Page 1 of 1
Who is online
In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 23 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science