triethylamine analysis
Posted: Mon Aug 30, 2010 10:44 pm
by martinamarie
Hi, can anyone tell me if triethylamine can be analysed by HPLC by forming a salt and detecting with UV detection, or would the analysis have to be with RI detection. Or is the only way to analyse triethylamine by GC
Tina
Posted: Mon Aug 30, 2010 11:23 pm
by tom jupille
Any of the above should work. Do a Google search for "triethylamine" + "indirect uv" and you'll come up with several links using capillary electrophoresis, but the same approach should work with HPLC.
A conductivity detector with a cation exchange column works also (I've done it that way in the early 80s)
Posted: Mon Aug 30, 2010 11:36 pm
by martinamarie
Thanks Tom, I'm on it
Tina
Posted: Tue Aug 31, 2010 8:43 am
by HW Mueller
Tina, amines have considerable absorption at or even above 200nm. Triethylamine has a max around 210nm.