What are the basic guidelines for using constant flow v/s constant pressure on an Agilent 6890?

My rule of thumb is easy. Unless I need to follow a method that specifies constant pressure, I use constant flow. I try to run columns close to the optimum linear velocity - or just a bit above that. If the low rate drops - as happens with temperture programs and contant head pressure - the linear velocity may drop below optimum. And the van Deemter curve is much less forgiving at low flow than high flow.

I agree, constant flow is better.

yeah

I worked in one lab that ran a LOT of methods in constant pressure. Problem was, they used a lot of 100 meter capillary columns and temperature program methods. With constant flow, the head pressure has to increase as the temperture increases. You run into a limit if you have a very restrictive column, like the long ones.

Either one will work. There have been many, many samples run in constant pressure for many years before electronic flow controllers were developed.
5 posts Page 1 of 1

Who is online

In total there are 58 users online :: 1 registered, 0 hidden and 57 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 57 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry