Indirect UV detection for analysis of choline
Posted: Fri Aug 27, 2010 10:58 am
by chromeleon
I'm using naphthalene sulfonic acid as my UV-active additive, 0.0008M. running this at 100%, 1 ml/min.
the baseline is very noisy when I approach ug/ml levels. I'm using a column heater at 40C and filtering my mobile phase.
can anyone suggest anything please?
also I've run choline and DMAE samples at 1; 1:10; 1:100 and 1:1000 - I'm getting peaks (negative) but the areas don't make sense, i.e. I'm not seeing the expected changes in magnitude.
if anyone has any experience of this LC mode I'd really appreciate your suggestions, thanks
Posted: Wed Sep 01, 2010 8:03 am
by Csaba
The background concentration of the UV-active additive must be much higher than the highest analyte concentration to be detected, at least 10 times higher, preferably 100 times. Indirect detection usually has a narrow linear range, in many cases the linear range is just one order of magnitude. I hope this helps.
Posted: Fri Sep 03, 2010 10:24 am
by chromeleon
why?
Posted: Fri Sep 03, 2010 12:17 pm
by carls
If memory serves me indirect UV is performed using a UV active additive with the same charge as the analyte. The measurement principle is based on maintaining solution electro-neutrality so the analyte displaces the UV active species having the same charge. This displacement is what gives the decrease (negative peak) in detector response.
This is from memory so hopefully someone will comment on the accuracy (or lack thereof) of this.
Posted: Fri Sep 03, 2010 1:39 pm
by chromeleon
carls, read my post again. I'm expecting negative peaks - the problem is magnitude. I was hoping to see linear relationship, i.e., the peak area is 10 times higher when the conc. is 10 times higher...
is there no-one at all on this forum with significant ion-pair experience???
Posted: Fri Sep 03, 2010 2:38 pm
by carls
perhaps the following paper will help:
http://www.mdpi.com/1420-3049/10/9/1179/pdf
Posted: Tue Sep 07, 2010 8:58 pm
by hajdaei
We tried an indirect UV method recently. We used 0.3 g/L benzoic acid in the mobile phase.
It did not work for us. Never really did figure out why.
Are there any inside tricks to getting this type of approach to work.
Posted: Wed Sep 08, 2010 7:00 am
by Mattias
Although I am not familiar with indirect UV, some general rules about UV detectors should still apply.
I guess the absorbance of your mobile phase is so high that almost no light passes through the cell. If you autozero your detector with water and then run your mobile phase through, you can then see what the absorbance of your mobile phase is. If it is above (approx) 1.5 AU, you will run into trouble. A lot of noise and non-linear behaviour.
Posted: Sat Sep 11, 2010 4:43 pm
by carls
chromeleon - are you still following this?
The system you describe is more than simple ion pair chromatography or simple indirect UV detection.
Choline carries a permanent positive charge and your ion pairing agent (naphthalene sulfonic acid) is also your indirect UV visualization agent?
The article I posted above mentions such a case (competitive adsorption?) but this, in my experience, is not a common approach.