The tiny peak area puzzled me - presumably Empower generates smaller numbers for peak areas than Chemstation - they are arbitrary units in any case.
If you have the peak in air blanks than it is either a) in the air, b) carryover, c) in the carrier gas, d) in the syringe.
What happens with a no-injection blank ? if you see the peak it is either carryover or in the carrier gas.
What happens if you do a series of no-injection blanks. If the peak gets smaller it is carryover, if it stays the same it is in the carrier gas.
You have changed the syringe and still see the peak, but the syringe might be getting contaminated by the purge gas, do you have a charcoal scrubber in that line ?
How are you cleaning your carrier gas, and if you have scrubbers when were they last changed ?
Try aninjection form a vial that has been flushed with clean nitrogen to elimiante air contamination - a little of what is in the air will be stuck to the inside of the vial of course.
You may need to change the split line filter.
Peter