Page 1 of 1
How to check the detector is working?
Posted: Mon Aug 23, 2010 12:41 pm
by sanatsmishra
Hello everyone,
I have setup an old HPLC instrument (was not used for 5-6 months) and am new to it. I wanted to check whether the FL detector is working or not I tried few assay which involve derivatization process, however I was not able to get the peaks. Is there any simple and convenient method available ?
Please let me know.
Thanks in advance.
Posted: Mon Aug 23, 2010 2:57 pm
by Schmitty
If you want to isolate the detector from the rest of the instrument and take all of the guess-work out of the sample preparation by removing the samples and standards entirely, what would you need in order to produce a (fluorescent) response?
Schmitty
Posted: Mon Aug 23, 2010 3:43 pm
by HW Mueller
Check the manual on how to do a Raman spectrum of water.
Posted: Mon Aug 23, 2010 9:51 pm
by tom jupille
Or, remove the column and do a direct injection of tonic water (contains quinine).
http://omlc.ogi.edu/spectra/PhotochemCA ... O4%29.html
You can dispose of the excess tonic water by mixing with an appropriate quantity of gin :wink:
Posted: Tue Aug 24, 2010 8:57 pm
by Schmitty
Yeah, that's what I was getting at...but I didn't want to just say do XXX. I think this is the test we did in college for one of our instrumental analysis classes to determine detector operation.