Page 1 of 2
Phosphatidyl Serine
Posted: Wed Jan 26, 2005 8:45 am
by syx
Dear Chromatographers,
Could anyone share the information about RP-HPLC determination of phosphatidyl serine? We only get some method using ELSD and TLC. We do not have ELSD.
Best regards,
SYX
Posted: Wed Jan 26, 2005 8:41 pm
by maris
The said compound has very "weak" chromophors,
what is the concentration range of it in your samples ?
For low concentrations UV detection will be useless
Which other types of HPLC detectors do you have?
Posted: Wed Jan 26, 2005 9:10 pm
by Kostas Petritis
CLND or MS would work for this...
Posted: Thu Jan 27, 2005 7:24 am
by syx
Bad news for me ...
I just have DAD and RID.
Is it possible to use derivatization mode?
Posted: Thu Jan 27, 2005 3:04 pm
by Kostas Petritis
RID should also work although the method won't be sensitive...
Posted: Thu Jan 27, 2005 3:56 pm
by Ryans17
There's a new detector out called the Corona CAD which may be a solution for you too? It is also a evaporative detector but involves charging.
Posted: Fri Jan 28, 2005 12:46 am
by syx
hmm... i think i should try RID first as Mr. Kostas said. But I will consider the other option if RID won't be run.
Otherway, I will try TLC-Densitometric method with iodine vapor derivatization.
Posted: Fri Jan 28, 2005 10:43 pm
by Einar Ponten
Let me be cocky. I'll eat my hat (or any suggested) if this compound does not have retention on the ZIC®-HILIC column.
However, as pointed out above the detection constitutes a problem and you need to find a suitable detector like ELSD, ECD or MS (or an old fashioned colorimetric procedure for post-column detection)
Posted: Mon Jan 31, 2005 12:53 am
by syx
... that's the problem. Have any idea for last option (colorimetric post-column detection)?
I need any reference about this technique. I never use it before and feel unfamiliar.
Posted: Mon Jan 31, 2005 5:23 am
by Uwe Neue
For options on derivatization reactions, look at the book by Lunn and Helwig. For options how to do it best, look for "knitted tubes". I believe that SeQuant can help you with the hardware. For the principles of the technique, you can ask me.
Posted: Wed Feb 02, 2005 1:32 am
by syx
Dear Mr. Neue,
I will search and read the book first, then ask you anything that I do not understand to you. Thank you.
Posted: Tue Feb 08, 2005 10:01 pm
by Mark Tracy
If you are in the mood for something experimental...
Phosphatidyl serine and phosphatidyl ethanolamine both have primary amino groups. These primary amines will react with ortho-phthalaldehyde and a mercaptan under mildly alkaline (pH 9-10) conditions. The derivative has absorbtion maxima at 330 and 230 nm and fluorescence emission at 465 nm. Please don't use mercaptoethanol; there are many superior alternatives. You could do this either by automated pre-column derivatization or by post-column reaction.
On the other hand, if you are going to be doing this frequently, maybe you should just write a strong case to your management for an ELSD.
Posted: Tue Feb 15, 2005 2:37 pm
by Klaus
Since we are taking about phospholipids:
What kind of phosphatidylserine are you trying to measure? Native or synthetic? (i.e. does it contain unsaturated fatty acids?)
What is the goal of your method?
Posted: Tue Feb 15, 2005 8:37 pm
by Einar Ponten
I agree with Mark. If you need an inexpensive, sensitive and quite robust detection then post-column reaction with OPA is an interesting alternative.
However, as pointed out above if you intend to run numerous of samples then an investment in a new detector will pay off.
Posted: Wed Feb 16, 2005 8:13 am
by syx
Mr. Klaus, we bought phosphatidyl serine derived from soy bean. We want to determine phosphatidyl serine in tablets.
I will do the experiment with OPA and mercaptan. What kind of mercaptan that is most recommended to try?