All in all I am happy with the results that I have been getting with the Atlantis column. However, I have noticed something that I find a little odd: this column seems to be going through something very similar to traditional column dewetting, but it is happening under 100% organic conditions.
Some of the methods that I am migrating to the Atlantis column have a 100% ACN wash at the end of the run. I have noticed that I have to let a 4.6 mm x 150 mm column equilibrate for about 20 minuets at a flow of 1.6 mL/min in order for me not to experience analyte retention time shifts in subsequent injections. Clearly, this retention time is excessive given flow and column dimensions.
I have found that if I change the wash step to 75:25 ACN:Water, then normal column equilibration times apply.
Initial mobile phase is 50:12:38 Water:MeOH:ACN.
I am curious to know if anyone else has observed this or, more importantly, if anyone can tell me what is causing this? Seems I have the problem sorted, but I am very curious as to the "why".