-
- Posts: 12
- Joined: Fri Dec 02, 2011 9:55 pm
We are looking to assay zinc phosphide, a rodenticide, in various biological matrices. Our current method involves adding an organic solvent (we use p-xylene) and sulfuric acid which reacts with the analyte and gives off phosphine gas. This takes place in a closed vial. The samples are then tumbled for 2 hours during which the phosphine gas is dissolved into the xylene layer. The xylene layer is then analyzed on a GC with an NPD and the phosphine is detected. We make a calibration curve beforehand with purified zinc phosphide and back calculate the amount of zinc phosphide in the starting matrix.
We have a lot of reproducibility problems with precision. We believe the instrument is operating correctly, and we have checked the homogeneity of the samples, which we don't think is the issue either. The samples sometimes degas when filling the GC vials.
Would it make more sense to just do this all by static headspace analysis? Skip the xylene, let the reaction occur in the headspace vial, and inject the gas? Phosphine is not water soluble so it would all enter the headspace. If so, what's an older model of HSA to use? We've found Varian Genesis and Tekmar 7000 units for sale.. are these identical systems? How about the HP 7694? Any experiences? We had issues at an older lab with the 7694 not dropping the vials into the oven correctly.
Thanks for your help.