'dancing' RT
Posted: Thu Aug 19, 2010 2:28 am
hi all,
I'm having an interesting situation here. I was given 3 samples of a dye from different manufactures to check the purity by UV. The dye is IR-783 (more info: http://www.sigmaaldrich.com/catalog/Pro ... KEY&F=SPEC )
I used a very long gradient (had plenty of time) 5-95 ACN:MeOH 1:1 in H2O (0.1M H3PO4) on a C18 column. I found that the RT of the 3 samples were different (up to 0.5 min). I used a different column, different gradient, different organic phase (ACN alone). Still, all of them coming at different RT.
So, I moved to a different instrument (which I just used for another project with no problems). Still, same results.
So, moved to another instrument (have plenty of them
. Check the full UV-vis spectrum and they are identical (which doesn't mean the structures are identical). Then, I ran 2 consecutive runs of the same sample and found that the RT were different. Took a vial from some old project and did 2 consecutive runs and got 2 perfect identical chromatograms.
Now I have to conclude that the IR-783 has an erratic behavior? Does anybody seen this type of behavior before?
BTW, although the manufactures claim "90% dye", I don't see any other peak on the UV-vis range. One interesting difference in the manufacture specs: 2 of them have melting point: 209-212C, the other one 256-259C. May be they are different crystallographic forms?, although there is no observable difference in the powder aspect. Anyway, a different crystallographic form shouldn't matter when they are in solution, right?
I really appreciate your thoughts
thanks,
Sergio
thanks,
Sergio